    Test Method:  CPSC-CH-E1001-08

    Standard Operating Procedure for Determining Total Lead (Pb) in  
    Children’s Metal Products (Including Children’s Metal Jewelry)  
    December 4, 2008
    This document provides detailed information on test methods that will be used by the U.S.
    Consumer Product Safety Commission’s (CPSC) testing laboratory (LSC) in the analysis of
    children’s metal products for lead (Pb) content.  The methods described determine the total lead
    content of metal items such as, but not limited to, children’s metal jewelry.   

    CPSC staff has concluded that these test methods are sufficient to determine lead content in most
    metal and metal alloy consumer product component parts.  Adjustments may be necessary for
    products made from certain metal alloys, and should be based on sound chemistry and materials
    science knowledge as well as appropriate acids for the alloy matrix.  Test methods selected
    should be those that best demonstrate that they can achieve total digestion of the sample material
    being analyzed.  Test methods should not be used if they are inconsistent with the specified
    application of the test method, or do not demonstrate the best performance or proficiency for
    achieving total digestion of the sample material.   

    The methods apply to metal and metal alloy components and have not been recommended by
    CPSC staff for other materials that will not dissolve under the conditions specified below or
    when combined with the specified acid(s) result in an inappropriate combination of materials that
    would be inconsistent with safe laboratory practices.

    The general approach is to grind any accessible component part of a sample to a powder; digest
    completely in a combination of hot, concentrated nitric and hydrochloric acids; and analyze by
    Inductively Coupled Plasma – Optical Emission Spectroscopy (ICP-OES).   

    Other analytical methods such as Inductively Coupled Plasma – Mass Spectrometry (ICP-MS)
    and Flame Atomic Absorption Spectroscopy (FLAA) and Graphite Furnace Atomic Absorption
    Spectroscopy (GFAA) may be used under appropriate conditions as an alternative to ICP-OES
    using applicable, recognized analytical techniques for the alternative analytical method.


    -2-
    Definitions

    1.  Sample – An individual consumer product or a group of identical consumer products
    from a batch to be tested.
    2.  Component Part – Individual sub-unit within the total sample.  An item such as a bracelet
    may be broken into its component parts such as a bead, a hook, and a pendant, with those
    component parts individually analyzed.
    3.  Instrument Detection Limit (IDL) – 3 times the standard deviation of 10 replicate
    measurements of reagent blank.  The IDL for Pb on the Inductively Coupled Plasma
    Optical Emission Spectrometer (ICP-OES) currently used by LSC is 0.01 ppm.
    4.  Method Detection Limit (MDL) – Reagent blank fortified with 2-3 times the IDL.  Seven
    replicate measurements are made.  Calculate the MDL as follows:  MDL = (t) x (S),
    where t = 3.14 (99% confidence level for 7 replicates) and S = standard deviation.  The
    MDL determined based on current LSC staff practices and equipment for Pb is 0.01 ppm.  
    5.  Laboratory Reagent Blank (LRB) – An aliquot of the digestion reagents that is treated
    exactly as a sample including exposure to glassware, digestion media, apparatus and
    conditions used for a particular Pb test, but with no added metal sample.  LRB data are
    used to assess contamination from the laboratory environment.
    6.  Calibration Blank – Deionized water acidified with nitric acid (3 ml concentrated nitric
    acid diluted to 100 ml with deionized water).
    7.  Stock Standard Solution – 1000 ppm solution of Pb (in nitric acid) purchased from
    reputable commercial source, used to prepare calibration standards.  Replace before
    expiration date.
    8.  Calibration Standard – Solutions containing 0 to 25 ppm of Pb in 3% nitric acid matrix
    are used.  A minimum of 4 calibration standards are used.  Calibration standards should
    be prepared weekly from the stock solution.
    9.  Quality Control Sample (QCS) – Solutions containing Pb that are used to evaluate the
    performance of the analytical instrument system.  QCSs are obtained from a source
    external to the laboratory and are not made from the Stock Standard Solution.
    10. Certified Reference Material (CRM) – CRMs are materials with similar matrices as test
    samples with known lead levels.  The CRMs are used to verify digestion and analysis
    methods.  For example, standard reference materials (SRMs) are CRMs that are available
    from the National Institute of Standards and Technology (NIST), such as those listed in
    the Equipment and Supplies section below.  

    Equipment and Supplies:  The materials used for sampling and analyses are as follows:

    1.  Nitric Acid, Trace Metal Grade
    2.  Hydrochloric Acid, Trace Metal Grade
    3.  Digestion Vessels, 50ml (glass or plastic suitable for the digestion conditions)
    4.  Hot Block Digester (such as Environmental Express HotBlock?)
    5.  Metal Cutters
    6.  Distilled Water
    7.  Microwave Digestion Apparatus
    8.  Rotary Grinder (such as Dremel® Tool, with carbide burr grinders)
    9.  Orbital Shaker -3-
    10. CRMs for Leaded Metals (such as NIST SRM 53e, 54d, 1129, etc.)
    11. Internal Standard (such as yttrium, from a stock standard solution of that element,
    appropriate to the instrument parameters of the ICP used for the analysis)


    I.  Total Pb in Metals - Digestion of Sample

    When preparing a sample, the laboratory should make every effort to assure that the aliquot
    removed from a component part of a sample is representative of the component to be tested,
    and is free of contamination.

    Each unique, accessible component part from a sample must be analyzed for total Pb content.  
    LSC uses either of two methods for determining lead content in metal items.

    One method used by LSC is a hot block digestion, based on methodology found in Canada
    Product Safety Bureau Method C-02.4.
    1
      This method is equivalent to the procedure referred
    to as I. Screening Test for Total Pb Analysis given in the CPSC Standard Operating
    Procedure for Determining Lead (Pb) and Its Availability in Children’s Metal Jewelry, dated
    February 3, 2005 (http://www.cpsc.gov/businfo/pbjeweltest.pdf).  Note that the second part
    of the referenced procedure for determining acid extractability of Pb is not applicable for the
    determination of total lead content.

    The other method used by LSC is a microwave digestion method based on EPA method
    3051A2
    (http://www.epa.gov/epawaste/hazard/testmethods/sw846/pdfs/3051a.pdf) with
    modifications as described below.   

    Modifications of these methods may be necessary for certain alloys, taking into account
    appropriate acids for digestion.  Certified reference materials that closely match the alloy of
    the tested product should be used to verify the accuracy of the digestion and analysis
    methods.
  • TA的每日心情
    开心
    4 天前
  • 签到天数: 35 天

    [LV.5]常住居民I

     楼主| 发表于 2009-2-21 23:35:00 | 显示全部楼层
    SAMPLE PREPARATION METHODS
    Either of two methods outlined below have been found by LSC staff to be suitable for
    preparation of metal samples within the scope of this document for subsequent elemental
    analysis.

      A.  Hot Block Method
    1.  If the item is coated with paint or a similar surface coating, the coating should be
    removed and analyzed separately from the base metal for lead content, as described in
    the CPSC Standard Operating Procedure for Determining Lead (Pb) in Paint
    (http://www.cpsc.gov/businfo/leadsop.pdf).  Care should be taken to remove as little of
    the substrate metal as possible.
    2.  Weigh out a 30-100mg aliquot of a component part in a labeled 50ml digestion vessel.
    Component parts of children’s products including metal jewelry items generally weigh
                                                    
    1
    Determination of Total Lead in Metallic Consumer Products
    2
    Microwave Assisted Acid Digestion of Sediments, Sludges, Soils, and Oils -4-
    several grams or more, and an aliquot (with no paint or similar surface coating, but
    including any electroplated coating which is considered to be part of the substrate) will
    have to be obtained.  Samples should be cut or ground into many small pieces to
    increase the dissolution rate.  If a grinding apparatus (such as a rotary grinding tool with
    disposable grinding bits) is used, then any contaminated parts must be thoroughly
    cleaned or disposed of between uses to prevent cross-contamination.  Record the actual
    weight of the aliquot of the ground up component part to the nearest 0.1mg.
    3.  In a chemical fume hood, add 8ml of concentrated nitric acid to each beaker and
    evaporate to approximately 3ml on a hot block digester at 105±3°C.
    4.  After cooling to below 50°C, add 2ml of concentrated hydrochloric acid and stir.
    5.  Dilute with distilled water, washing the side of the beaker, to 20ml.
    6.  Warm up the solution to at least 60°C without boiling and gently agitate on an orbital
    shaker or with a stirrer or shaker bath for a minimum of 4 hours.
    7.  Transfer quantitatively into a 50ml volumetric flask or disposable volumetric digestion
    cups and dilute to 50ml with distilled water.
    8.  Dilute samples so that Pb results are within the calibration range of the instrument.  
    Generally a 1:50 dilution is sufficient.

      B.  Microwave Method
    1.  If the item is coated with paint or a similar surface coating, the coating should be
    removed and analyzed separately from the base metal for lead content, as described in
    the CPSC Standard Operating Procedure for Determining Lead (Pb) in Paint
    (http://www.cpsc.gov/businfo/leadsop.pdf).  Care should be taken to remove as little of
    the substrate metal as possible.
    2.  Weigh out a 30-100mg piece of metal item into an appropriate microwave vessel
    equipped with a controlled pressure relief mechanism.  Component parts of children’s
    products including metal jewelry items generally weigh several grams or more, and an
    aliquot (with no paint or similar surface coating, but including any electroplated coating
    which is considered to be part of the substrate) will have to be obtained.  Samples should
    be cut or ground into many small pieces to increase the dissolution rate.  If a grinding
    apparatus (such as a rotary grinding tool with disposable grinding bits) is used, then any
    contaminated parts must be thoroughly cleaned or disposed of between uses to prevent
    cross-contamination.  Record the actual weight of the aliquot of ground up component
    part to the nearest 0.1mg.
    3.  In a chemical fume hood, add 4.5ml of concentrated nitric acid and 1.5ml of
    concentrated hydrochloric acid to each vessel.  Wait for the initial reaction of acid and
    sample at room temperature to be complete (to the point of no obvious fuming or
    bubbling) before sealing vessels.  Seal vessels in accordance with the microwave
    digester manufacturer’s directions.
    4.  The microwave method should involve increasing the temperature of each sample to
    175±5°C in approximately 5.5 minutes, and holding at 175±5°C for 4.5 minutes.
    5.  Allow the samples to cool for a minimum of 5 minutes before removal of the vessels
    from the microwave.  Vent the microwave vessels in fume hood before uncapping.
    6.  Quantitatively transfer the sample to a 50ml volumetric flask or disposable volumetric
    digestion cups.  Dilute to 50ml with deionized water.
    -5-
    II.   Total Pb in Metals - Analysis of Sample Using ICP Method

    Analyze diluted samples for Pb concentration using an ICP spectrometer (or Atomic
    Absorption spectrometer).  Analysis procedures for ICP-OES and FLAA and GFAA are
    based on the methodology in ASTM E1613-043
    .  ICP-MS may also be employed with
    appropriate procedures, such as EPA 6020A4
    .  Calculate total lead concentration in the
    component part from that of the diluted sample, accounting for all dilution and reported as
    percent by weight of the component part itself.

    ICP Operating Procedures and Quality Control Measures

    Analysis
    1.  Ignite plasma.  Perform wavelength calibration or torch alignments per instrument
    manufacturer’s recommendations.
    2.  Allow the instrument to become thermally stable before beginning.   
    3.  Ensure the following element and wavelength are selected in analytical method:
    a.  Pb 220.353
    b.  One other Pb line, such as Pb 217.00, to ensure spectral interferences are not
    occurring during analysis.
    4.  An internal standard such as 2 µg/ml yttrium is used.
    5.  Perform calibration using Calibration Blank and at least 3 standards appropriate to the
    concentration of the analyte and instrument, such as 0.25, 0.5, 1.0 and 5.0 µg/ml.  
    Calibration shall be performed a minimum of once a day when used for analysis, or each
    time the instrument is set up.  Results for each standard shall be within 5% of the true
    value.  If the values do not fall within this range, recalibration is necessary.
    6.  Analyze the QCS immediately after the calibration.  The analyzed value of Pb should be
    within ±10% of the expected value.  If the Pb value is outside the ±10% limit,
    recalibration is required.
    a.  At least one LRB must be analyzed with each sample set.  If the Pb value exceeds
    3 times the MDL, then laboratory or reagent contamination should be expected.  
    The source of the contamination should be identified and resolved before
    continuing analyses.  The LRBs should be the same acid concentrations as added
    to the sample, and should be taken through the same digestion procedure but
    without added sample.
    7.  At least one certified reference material (CRM) should be analyzed with each batch of
    samples.  The CRM should be similar material as the test specimen with a known amount
    of Pb.  Analyte recoveries should be within ±15% of expected values.  If recoveries are
    outside this limit, the source of the problem should be identified and resolved before
    continuing analyses.
    8.   Dilute any samples that have Pb values exceeding 1.5 times the high calibration
    standard, and reanalyze.

                                                    
    3
    Standard Test Method for Determination of Lead by Inductively Coupled Plasma Atomic Emission Spectrometry
    (ICP-AES), Flame Atomic Absorption Spectrometry (FAAS), or Graphite Furnace Atomic Absorption Spectrometry
    (GFAAS) Techniques
    4
    Inductively Coupled Plasma-Mass Spectrometry -6-

    Calculations and Results Reported

    Results for the Pb test methods are calculated and reported as follows:

    Total Percentage Pb:    % Pb (wt./wt.) =  [(C x D) / (W x 1000 µg/mg)] x 100% where:

    C  = concentration of Pb detected (in units of µg/ml)
    D  = dilution factor (in ml units)
    W = weight of aliquot digested (in mg units)

          Example:

    A 50 mg aliquot of a component part was digested and diluted to 1000 ml in order to analyze
    by ICP.  The ICP analysis found 20 µg/ml of lead in the solution, which showed that the
    original component part contained 40% Pb by weight.
  • TA的每日心情
    开心
    4 天前
  • 签到天数: 35 天

    [LV.5]常住居民I

     楼主| 发表于 2009-2-21 23:43:00 | 显示全部楼层

    (C)
    (D)



    (W)


    Component
    Concentration
    Detected on ICP
    (µg/ml Pb)
    Dilution
    Factor(ml)
    Total Pb
    (µg)

    Sample wt. (mg)
    % Pb
    Pendant 1
    20
    1000
    20,000
    50
    40

    http://www.cpsc.gov/about/cpsia/CPSC-CH-E1001-08.pdf